关于设备作业平台对地系统电阻测量方法

Hello, Johnk:

I would like to ask what the compliant method for measuring the ground system resistance of the automated equipment operation platform is? Currently, we have two measurement methods internally as follows: 1. Use a heavy hammer electrode, with one test line connected to a heavy hammer electrode placed on the operation platform and the other test line connected to the ground system to test the ground resistance of the operation platform. 2. Use alligator clips, with one test line connected to an alligator clip touching the operation platform surface and the other test line connected to the ground system to test the ground resistance of the operation platform. I believe that the ground resistance test of the operation platform should use a heavy hammer electrode to test the ground system resistance (because the heavy hammer electrode simulates the large-area contact of objects with fixtures in actual use, fully testing the full-path resistance from “fixture surface → material → ground terminal → earth”, and truly reflecting the static discharge capability). Which of the two methods is more recommended as the standard measurement method?


Either method is fine. Using the five pound electrode will have a low limit on the resistance measurement while the alligator clip will be lower.

First of all, thank you for your answer. Currently, we have a problem. When we measure the system resistance to ground using a five-pound electrode, it is qualified. However, when we measure the resistance to ground using the second alligator clip in the above picture, it is unqualified (exceeding 10 to the power of 9). I think it is caused by the influence of contact area and weight. So, I want to use a heavy hammer electrode to measure the resistance to ground. Is this more reasonable?

Most likely the plate may be anodized which would make the alligator clip hard to have a connection. The 5 pound electrode would be fine to measure that.

Yes, the surfaces of these fixture platforms are all made of anodized material. Thank you very much for your answer. Thank you!

Hello ,
Greeting of the day from India to all the members on this platform.

As you have shown in the first photo and as suggested by our JK Sir, it is ideal and correct to use the 5 lb probe which develops perfect contact.

Alternatively, you may use a flat SS rectangular plate kept on the bed fitting to the said area and check with the 5 lbs probe.